Constant flow v/s Constant pressure
What are the basic guidelines for using constant flow v/s constant pressure on an Agilent 6890?
My rule of thumb is easy. Unless I need to follow a method that specifies constant pressure, I use constant flow. I try to run columns close to the optimum linear velocity - or just a bit above that. If the low rate drops - as happens with temperture programs and contant head pressure - the linear velocity may drop below optimum. And the van Deemter curve is much less forgiving at low flow than high flow.
I agree, constant flow is better.
yeah
I worked in one lab that ran a LOT of methods in constant pressure. Problem was, they used a lot of 100 meter capillary columns and temperature program methods. With constant flow, the head pressure has to increase as the temperture increases. You run into a limit if you have a very restrictive column, like the long ones.
Either one will work. There have been many, many samples run in constant pressure for many years before electronic flow controllers were developed.
Users browsing this forum: Semrush [Bot] and 2 guests