Connecting Varian 9050 UV-VIS detector to Ultimate3000HPLC
I need connection of Varian 9050 detector to Ultimate HPLC system via UCI-50 Dionex Interface to be collecting data under Chromeleon.
This is virtually was connected from HPLC side, but I do not have
cord ending to the Varian detector. From where I can buy such ending and
what is the english name of such part.
Kazimierz
Do you have the cables that came with the UCI-50?
Most detectors have an analog out connection to record the signal and also an auto zero input connection.
Alex
Dear Alex,
The servicemen have attached two additional data transfer cords
(one Shimadzu RF-551 spectrofluorometric detector as well as to
Varian 9050 UV/VIS Detector - using CH1 and CH2 inputs respectively.
Ultimate3000 is connected using one USB input/output cord
The Shimadzu working very well because it have possible outside ways
for connection, without the ending: black-is shield: pink-is analog high+;
and gray- is analog low-. The both cords was taken from original
ULTIMATE3000 service spare parts shipped with HPLC: QC 8911 0001,
P/n 6000 1004 but both don't have any endings.
Varian 9050 UV/VIS detector have both identical outputs 1 or 10 V
dedicated for external DATA SYSTEM OUTPUT and for RECORDER OUTPUT,
respectively.
I need one ending of 3 fibers to be connect detector to Chromeleon.
sincerely
Kazimierz
Sounds like you need a cable from Varian that would plug into the detector output.
That cable would be spliced onto the Dionex UCI cable.