Chiral separation problem
A student at our institute is attempting to separate by chiral HPLC the (4',4'-R,R) and (4',4'-S,S) enantiomers of 2,3-di-(1'-(cyclohex-1'-en-4'-methanol)-yl)-buta-1,3-diene in order to determine the enantiomeric outcome of a stereoselective reaction. A very slight separation of the racemate is observed on a ChiralCel OJ-H 250mm x 4.6mm at 20 Celsius with 3% isopropanol in hexane at 0.5ml/min but this is insufficient for determining the enantiomeric excess. We have been unable to improve on this with the other chiral stationary HPLC phases available to us (Chiralcel OD-H, Chiralpak AS-H, Chirobiotic-R, Whelk O-1, Chiradex).
In contrast, the (4',4'-R,R) and (4',4'-S,S) enantiomers of the compound lacking the 1,3-butadiene moiety ( separates to baseline under the above conditions. Any suggestions as to how solve the separation problem of the enantiomers of the compound with the 1,3-butadiene moiety would be greatly appreciated. I can e-mail the structures of the compounds. Thank you in advance for your time.
for this diene, or the part# from the vendor? so i may get some to try. i am in US.
thanks,
You might like to try the IA, IB or IC columns from Daicel. They offer the chance to use non-standard solvents (THF, DCM, EtOAc etc) and therefore bring alternative selectivities into play.
Anthony.
what is the requirement on hardware to run these types of exotic solvents?
as I know parts made of peek will swell against some solvents, like DCM. not sure of teflon material either.
thanks,
I have only used them on systems plumbed with stainless steel. You are correct that you need to ensure system compability before attempting to use these solvents.
In total there are 26 users online :: 2 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Google [Bot] and 24 guests
