Volumetric flow meter
I'm looking for a volumetric gas flow meter with the the following characteristics:
- Accururacy better than 2%, 1% prefered
- Large flow range, about 2 to 1000 ml/min.
The volumetric meters I have found so far are either too inaccurate or only takes samples of the flowrate.
Any tips or links?
Do you need a on-line flow meter with those specs ?
that is not impossible( but almost hardly, and not
conventionally) may be you need tell us your application
Daniel H. Sagarra
Field Application Engineer
Buenos Aires, Argentina
www.flowmeet.com
I came across a flow meter called Reciflow,
http://www.ecinstruments.com
From the spec you have presented, it seems it could be a good fit for you.
Is this the type of instrument you had in mind?
Thank you very much, Dieter!
It's more or less exactly what I'm looking for. Not only a calibration instrument for other equipment, as other piston meters I have seen, but it can also be used inline in my application. Great!!
In total there are 20 users online :: 1 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Users browsing this forum: Google [Bot] and 19 guests
