Dionex Pump Problem
I have this Dionex DX300 system with the GPM 4 solvent pump. I run carbonate buffer through an anion Dionex column at 1.2ml/min running a standard Anion analysis.
I was running it normally when suddenly the pressure dropped. I noticed the purge valve was also blocked. I could'nt draw any liquid from it.
I tried using the second solenoid from no. 2 position instead of no. 1 and still the same problem. So all 4 solenoids block and I cannot draw liquid from the purge valve, and of course I cannot also pump through the column either.
I also noticed that If i leave the instrument off for like 1-2 minutes and then turn it on again the problem persists, whereas if I leave it off for like 15 minutes or more it tends to work again for 5-6 minutes and then the problem come back again..
Anyone that has experienced similar symptoms or could explain this?
Sounds like electronics / power supply overheating. Check the exhaust fan to see if it's running. Fans are cheap if that's the problem. You can get one from FRY's.
C.Tony Vella Royal British Legion
WWW.HPLCworks.net
858.663 751
Arte et Marte