HPLC analysis of bupropion
It is my first time to post here. I hope I would find help. I have been investigating some xenobiotics for their inhibitory action on Liver CYP 450 enzymes. I am facing a problem in measuring hydroxybupropion as there is no peak detected by HPLC. I have used 0.4 mg/ml protein and final concentration of Bupropion 50 uM. The eluent I used was 75% 50 mM o-phosphoric acid buffer PH 3 : 25% acetonitrile. . Could you tell me how can I overcome such a problem and is it helpful if I use higher concentration of microsomes or Bupropion or both?
Thanks in Advance
:lol: Dear Manar_fawzy,
Being high UV activity, Bupropion it should give peak at 50 uM.
Initially u should have to setup concentration range by injecting Individual solutions.
if u still not getting peak at 50 uM concentration, u may play with Injection volume, mobilephase & UV detection.
Ignore if u already did that.
Mouli
Hi
Thanks for your help. I would give more details. This is reversed phase chromatography. I used in all my trials 20uM injection volume. UV wave length is 214. Device is Shimadzu, Column is Waters symmetry C18 5um,3.9x150mm. What do you think I should try?
What do you recommend should I make the mobile phase more or less polar?
Kind Regards
Manar
Unless you are working from a published method, the easiest approach would be to start with a pure standard and run a gradient from, say 5%to 95% organic. Ideally, you will see a single peak in there somewhere. Look at the % organic at the time that peak elutes and back off about 10% from that as a reasonable starting point for developing an isocratic separation. Fine-tune as necessary from there.
In total there are 49 users online :: 2 registered, 0 hidden and 47 guests (based on users active over the past 5 minutes)
Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 47 guests
