HPLC 1090 auto sampler problem
The auto sampler is not working due to low air pressure. If there is a certain value I have to check. Any advice will be appreciated. Faris
Is your house air supply or cylinder supplying adequate pressure to the autosampler? If so, you may have an internal air leak.
You have to be able to provide 5 bar on the inlet of the instrument. The easiest way is to have a compressor like we to in our lab.
Hp 1090 contains a pressure sensor mounted under the auto sampler table which show that the air pressure is ok or not. you should watch instrument condition in chem station panel. there is a text (airp) with yellow background sign if no pressure available and it goes green if the air pressure become adequate.
