Of course , I can't find the manual. I am getting an error code upon startup with continual beeping. The code is 5t. Can't find anything on a google search. The unit is about 10 yrs old. Thanks.
I recommend contacting Biotage. They now manufacture this equipment.
Could your code be St? That is when the desired temp is above that allowed when sensors are installed ( 60c).
That was it, someone changed the bath set temp to 70C. Thank you for your help!
Glad I could help.