random nosie in FID
What could be reason of random negative-positive noise in FID.? I Conditioned 3 hours and also change flame tip and tube collector. But no useful
Hi!
What gas flows do you set in FID?
Does your gas generators in proper condition?
All I ever need to know I'm learned in cloning vats.
Hi!
the GC is three channel and other channel are ok. anyway carrier gas is heluim, makup is nitrogen from plant, hydrogen from generator and air from plant. if problem was related to gas we should have some noise in other channel but only one have noise!
Could just be a detector that needs a regular cleaning maintenance.
mm
might not be related, but I occassionnaly find it suprising, that when operators change flame tip, they do not spend 30s to check for ferrule residues (from overtightning etc), that may cause similar random spikes, depending on desing I suppose.
finally found the problem. that was related to electronic FID board. thanks for all replyes
In total there are 77 users online :: 1 registered, 0 hidden and 76 guests (based on users active over the past 5 minutes)
Users browsing this forum: Ahrefs [Bot] and 76 guests