When I used to work in the lab, I used an OI 4660 for two years and had some serious problems with a few key compounds. Never got reproducible results for two compounds. Moved to Tekmar and got much better results immediately.
I changed the trap when I lost response.
Don
Don
Chem Sam- Which key compounds did you have problems with?
Don- How often were you changing the trap?
I was trying to remember which analytes - they were heavies - I remember one being napthalene being one. I know when we changed out two of the sample pathway lines, the problem went away. But it came back within 4 months.
In regards to how often you have to change the trap - this definitely depends on how many samples you analyze. It's my understanding that most people change it ever 8 mo to a year.
*****Commercial Here******
If you choose to look at alternatives, try http://www.estanalytical.com/Products/E ... Evolution/ . We will guarantee your results and give you any help necessary.
If you are only having problems with you heavies, it could be a trap issue, but it could also be a Cyclone mgmt issue that has developed an active site. If you cannot get the help you need, feel free to call our support line at 800-283-3510. They can help you free of charge over the phone.