HPLC Error!
Hello all,
after starting the run on schimadzu SCL 10Avp hplc i got an error as SIL: NO 6PV VALVE HOME. i got this problem first time. what shall i do please let me know if anyone has the right answer will be appreciated.
zulfaquar
Your autoinjector can't properly actuate the 6 port valve.
This error is usually caused by wear over time resulting in polishing of the mated surfaces. Once they are polished enough to allow for significantly more contact than with new components, the actuating motor can no longer overcome the friction between them, causing the error.
At the very least, a qualified person needs to open up the 6 port valve and inspect the rotor and stator. It would probably be best to go ahead and replace those parts.
Users browsing this forum: Google [Bot] and 11 guests
