Quattro Ultima issues
I am having trouble hooking up an HP1100 HPLC to the Quattro Ultima triple quad MS. As of now there is a relay cord running from the HP 1100 autosampler to the contact closure 1 of the MS, but I can't seem to get that connection to work. When I go to configure the MS inlet method it gives me an error of "Failed to open sicl communication session at address hpib7,21. SICL error: -No error. Does anyone know what's going on.
if you are still having trouble with the communications let me know your email address and I can send you a setup guide for the lc / software for agilent control with masslynx
I still am having issues with the connection but i think it is due to the drivers for the HPIB card are not in masslynx. I need the instillation cd or someone who has the cd for masslynx 4.0 so i can get the drivers off of it. If you know anyone who has this my email is narient_82@yahoo.com.
I also am running into this issue, although I know the system drivers for the HPIB interface are installed. I'd really appreciate any guidance anyone could give on this topic: jkhilmer@gmail.com
I found out that the issue was not the drivers or the Masslynx software but with the Agilent side. I was running an Agilent 1100 series LC and Chemstation software that didn't allow for the LC to be hooked up via HPIB. So you need to make sure that your version of LC software can handle the HPIB for your system. I was using a 82350B HPIB PCI ADAPTER CARD because it is the only one that Masslynx will recognize. But your software may be different and I would advise checking card compatibility.
I would like a copy of the guide sabbathpm@yahoo.com
In total there are 34 users online :: 1 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Users browsing this forum: Semrush [Bot] and 33 guests
