Internal Standard - Suggestions Appreciated
I'm looking for a simple, stable molecule to use as an internal standard (to compensate for solvent loss during sample preparation).
The chromatography will be approximately:
Mobile Phase: 50/50 v/v MeCN / water (possibly buffered)
Column: Waters BEH C18 (1.7um)
Detection: 250nm
Any pointers to a group of candidate compounds would be greatly appreciated
Mark
Years back, I remember seeing alkyl phenones suggested as internal standards. You stary with a mixture and then pick a chain length that gives reasonable retention.
Tom
I seem to remember the same and have, (many years ago), used Benzophenone or similar as an internal standard.
Many thanks for the suggestion, it's a great place to start
Mark
