OPA/MPA derivatizatio stability for amino acids
I would like to know if the OPA/MPA derivatization results in stable products or is there something more I can do to avoid secondary derivatization products.
Please let me know about your opinion on this and what is your experience with the OPA/MPA derivatization.
Thank you very much
As far I know the products are not quite stable, so the best is to derivatize just before use in a time controlled situation. At least then there is a compareable situation between samples
As good as I know problem is not with secondary derivatization but with decrease of signal strenght in time due to decomposition of fluorescent product into non-fluorescent one.
With right type of autosampler you can run injector program that would perform OPA/MPA derivatization for you just before injection. I know Agilent 1100/1200 can do this.
For stable derivatization look at fluoresceine (less sensitive as good as I know) or AccQ Tag from Waters.
Check out the Pico tag from Waters if you have a standard HPLC. AccQ tag is for UPLC instruments.
In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 17 guests