What is reasonable variation of pH measurements?
We are producing a liquid product that needs to under quite tight pH control. For this reason, the production adjust the pH and measure the pH in several in process controls. The product contains a buffer with appropriate pKa.
When the bulk is ready the pH is measured again, and a sample is sent to QC for verification (100 meters away). The problem is that there seem to be a high variation between measurements in production and in QC, about 0.1-0.2 pH units. Shouldn't it be better than that?
The pH meters are calibrated just before each measurement, and pH is checked with a known buffer.
Changes in measured pH in the same product but in different labs are quite common.
If your pH needs to be in a very narrow range you need to make sure that the pH measurement takes place
- at a very stable temperature
- use temperature compensation
- with a propper calibration (fresh buffer solutions)
- with a propperly mainained pH electrode (clean diaphragm, correct inner solution etc.)
A difference in measurements of 0.1 to 0.2 pH units is not very big.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Thanks Markus!
Neither QC nor the production used any temperature control, so we are installing that now!