Oxygen related peak
I am seeing a peak in my chromatogram that I've determined to be caused by an elevated oxygen level in the sample. I am trying to mitigate its impact by altering the amount of dissolved gases in the mobile phase and/or sample. The link suggests that a negative peak can be obtained if the mobile phase contains more dissolved oxygen than the sample. This doesn't make sense since the lack of something does not produce a peak directly. Is it more likely that the peak attributed to oxygen is actually a solvent peak with a higher concentration of dissolved oxygen?
For anyone with experience on this issue:
Is it easy to nullify an interference related to dissolved oxygen by 'balancing' the absorbtion of the sample and mobile phase?
http://www.shimadzu.com/an/hplc/support ... 37lab.html
Actually, the lack of something *does* show as a negative peak at the same retention time as the presence of that something. The phenomenon is called "vacancy chromatography" and has actually been exploited as an analytical technique.
That is a "distinction without a difference". :wink:
"Absorbance matching" *can* be done, but is a PITA to get those small differences to match up, so very few people do it. A better (easier) approach would be to tweak the selectivity so as to move your analyte peaks away from the interference.
Thanks Tom. I'll go with your suggestion. This situation revealed an unexpected source of interference. I've never seen this before. I'm sure it's been present but it hadn't created a problem.