Mobile phase ions
emily lee
I always expect some noise ions at the lower masses. To see whether the amount of background you are seeing is normal, it would be good if you could look at old data and see if historically those ions have always been there at those levels (even when the instrument was installed), or if they are new contaminants. I checked my lists of common background ions, and the only one I see from your listing is 146, acetonitrile trimer with sodium, (3ACN+Na)+.
To answer your question of how LC/MS can be useful for identifying compounds when background ions are present: every data system I have used has a background-subtraction feature, that can be used to remove the mobile phase background ions whether you are doing flow-injection or LC.
MG, thank you very much. Since I take over this instrument, most of the time I run MRM for quantification. I really don't have old data to compare with. Would you tell me, what ions do you usually observe? and in what intensity? Thanks again.
emily lee