Headspace cleaning from DMF
Hi Guys!
I have a question. After the analysis of residual solvents in DMF, doing analysis of residual solvents in the water. how to completely clean the headspace system from DMF? (DMF- N,N-Dimethylfromamide (Sorry)
GC-HS agilent 7890A-G1888
many thanks
Greetings!
Your headspace system it's automated? The analysis it's from gc/ms or gc/fid?
It's dimethylsulfoxide or dimethlyformamide ? Dimethylsulfoxide is commonly expressed by DMSO. Either way i think the procedure will be the same for both.
If it's automated , remove the syringe, rinse the syringe about 3 times with acetone and 3 times with methanol. I reccomend a five minutes ultrassonic bath in methanol too, if you have this equipment of course. Put the syringe in the automated headspace and purge it for 5 minutes with nitrogen or other (cheap) gas that you use, run a blank.
if it's not automated, rinse with the solvents listed above about 5 times each. Wait the syringe to get dry and run a blank.
If in both cases you see something indicating contamination, consider change the septum and liner too.
:D
t.b.
T.B.
Greetings!
It's dimethylsulfoxide or dimethlyformamide ? N,N-Dimethylsulfoxide is commonly expressed by DMSO. Either way i think the procedure will be the same for both.
:D
t.b.
Ops!
your equipment is this : http://www.chem.agilent.com/Library/bro ... 6491EN.pdf
right?
I have a CTC-pal system coupled with a 5975C-7890A Agilent, an automated headspace system pretty simple acctually. I don't know if the procedure i listed will be able to clean your system. I really don't know how your system works.
Sorry.... :( ....
T.B.
Ops!
right?
Sorry.... :( ....
thanks anyway :)))
DMF is miscible with water, just put a mL of water in 10 vials and then equilibrate them in the oven at like 110 for a few minutes and inject, set the other heated zones to elevated temp and your GC's inlet to split mode with a few hundred mL of split flow. Set the oven above 100 C too, so no water collects anywhere. You could even remove your column if you want, just install an inch of two of column with a column nut into the inlet so it holds pressure.
If you still see DMF peaks after cleaning you may need to run more vials but I would bet it will be gone. The syringe is more difficult to get to in a G1888 than a CTC sampler. There is also significant flow path and a 6 port valve. All of it is heated so a bit of steam should help purge everything out, but I wouldn't go taking anything apart unless I needed to.
If you still see DMF peaks after cleaning you may need to run more vials but I would bet it will be gone. The syringe is more difficult to get to in a G1888 than a CTC sampler. There is also significant flow path and a 6 port valve. All of it is heated so a bit u very of steam should help purge everything out, but I wouldn't go taking anything apart unless I needed to.
Thank you very much :))
Users browsing this forum: Google [Bot], Semrush [Bot] and 6 guests