RP HPLC method for Imidazole
12 posts Page 1 of 1
Hi!
I need to detect imidazole in my API. Its is the reagent used in last step of API synthesis. I need to quantitate it using RP HPLC method with C18 column along with other synthetic intermediates and API.
Only column I have is Higgins Phalanx C18 100x4.6 mm 5 micron.
The mobile phase I have for this method is 10mM Ammonium acetate (pH-5.0) and 0.1% Formic acid in ACN.
Will it retain imidazole for atleast rt of 3.0 min? Should I increase the pH of the mobile phase? or change my mobile phase?
Any help will be highly appreciated.
Retention time of at least 3 minutes? That's easy, just use a flow-rate of ~0.3mL/min with your 100x4.6mm column. Then it doesn't matter which mobile phase you use.
Sorry, could not resist to give that answer :twisted: . It's nonsense to aim for a certain retention time, the retention FACTOR is the value which will show you if you have enough retention.
That said, Imidazole is a quite polar base, therefore I'd suppose it will be hard to retain on plain C18. HILIC is probably a far better option. If you only have C18 available, an anionic ion-pairing reagent like SDS or some long-chain alkyl sulfonic acid might be necessary.
Mixed-mode cation exchange on Primesep 100 column will give you retention of K'>3.
Vlad Orlovsky
HELIX Chromatography
www.helixchrom.com
I agree, mixed mode is a good idea.
If you want to stick with your column just add 0.2% TFA or HFBA. No buffer just acid. That should retain your target. Let it equilibrate for at least 15 minutes.
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island
How does TFA will help? Should I just use 0.2% TFA in water as Mobile phase?
Hi bobby12,
The purpose of the TFA and/or the HFBA would be to serve as ion-pairing agents in this case. Imidazole is a weak base (pKa ~ 6.9), so the idea is to completely ionize it and pair it in its ionized form with tetrafluoroacetate or heptafluorobutyrate ion to help retain it on a hydrophobic stationary phase. If you're "stuck" with the Higgins C18 column this is a good thing to try. No reason to stick with ammonium acetate at pH 5 or so (as suggested immediately above, wisely), just go towards higher acidity (lower pH) in the aqueous portion of your eluent. Keep the ACN at some level...TFA and/or HFBA may be added to both the aqueous and organic portions of your eluent(s)...not sure if you're trying gradient(s) or are shooting for an isocratic separation. Probably you'll find you'll need a relatively high proportion of aqueous phase in the eluent (mix).
I agree with the others...HILIC, mixed-mode or even maybe phenyl phases may be better choices to retain imidazole. ScienceDirect (web site of Elseveir) lists a few methods worked out by others for imidazoles, you can see the abstracts for free if you'd like to take a look at them, there aren't too many to look at, though.
If TFA or HFBA don't allow for sufficient retention, I'd try then something like a pH 2.5 phosphate buffer with a alkylsulfonate salt as HPLCAddict suggests. I'd kind of be surprised if HFBA didn't give you enough retention, though.
MattM
Hi! all,
Thanks for helping me out with this. I was able to get retention and good peak shape at 3.3 min by using 20mM potassium phosphate monobasic with 10mM octane sulfonic acid as ion pairing agent. pH=7.0.
95%buffer :5% MeOH as my mobile phase with Phalax C18 column.flow rate 1.2ml/min ,220nm wavelength.
Really appreciate all the suggestion. But got another issue is baseline shift. I have to resolve my API with imidazole so increased the mobile phase from 5% MeOH to 80% ACN:MEOH (60:20) in next 15 min.
0min 95% A (buffer): 5% B (meoh)
5 min: 95% A (buffer): 5% B (meoh)
15 min: 20% A (buffer): 20% B (meoh):60% ACN.
20 min :20% A (buffer): 20% B (meoh):60% ACN
21 min: 95% A (buffer): 5% B (meoh
AT 220nm the baseline is shifted upwards after 6 min and does not go down. Chromatography is not looking great. I don't know how to post pic otherwise I would love to show chromatograms.
How can I fix it.?
Thanks again
Monika
Hi Bobby12,
Just general comments:
Seems to me that the pH where you're operating (7.0) for the phosphate with the Ion-Pairing agent octanesulfonate is really kind of high...better protonation of that second nitrogen on the imidazole ring would occur at pH values of 5 or less (that's why I suggested 2.5 above). Maybe Tom Jupille will see this thread and lend a hand with embedding chromatograms...he once was kind enough to tutor me on embedding quotes, but I didn't use the knowledge he gave and misplaced his e-mail note to me.
Anyhow, if the retention time for the imidazole-octanesulfonate "ion-pair" in this case is able to be repeated consistently run-after-run...well, that surprises me, and you know what is happening better than I do.
Undoubtedly part of the shift in the baseline is due to the "proportioning-in" of organic...you've got a ternary gradient going on, and the buffer proportion decreases fron 95 to 20%. Some shift in the baseline would be expected...however, the change in percent organic will also have a possibly ill-effect on the concentration of the octanesulfonate that becomes "attached" to the stationary phase. Generally that is why gradients are not desired for use with ion-pairing agents as the ion-pairing agent can be slightly removed from the stationary phase during the course of a gradient, possibly affecting retention times for analytes in subsequent runs, and certainly contributuing to a shift in the baseline of a given chromatogram. Remember also...please allow some time for the HPLC system to recover from 20 Buffer:20 MeOH:60 ACN back to 5 Buffer:95 MeOH--didn't see this in the gradient table you typed in, and I know that the "guideline" of 10 column-volumes of initial eluent composition isn't written in stone, but one minute may not really be enough...particularly with octanesulfonate in there.
Again, you see what happens run-to-run using the separation you're working on...just something to consider. One thing, though...if your analytes are "spread-out" as much as you're describing, with that column and those conditions, you wouldn't need a column anywhere near as long as the one you're using. (As HPLCAddict wisely notes, it's not about the retention time of the analyte, it's about the retention factor of the analyte, the analyte retention time divided by the "retention time of an unretained analyte," sometimes called tm...for the imidazole-ion-pair in your case, the value is ca. 3.5, which is adequate for quantitative work.) Probably though, if this column is what you've got and you can't change...well, it's a situation I've seen in my career. Just because I think things may have been unwise didn't mean I could do a whole lot to change them...a more polar phase may help retain imidazole even without an ion-pairing agent, and not hold quite so hard onto you're other analyte of interest that contains an imidazole functional group.
My hope is that somehow you'll get to change the stationary phase to one that is more polar (for this set of analytes)...well-covered phases with excellent endcapping such as the one you're using are great for bases like imidazole, but in my experience only if I used acidic eluents and ion-pairing or basic eluents with pH greater than the pKa values of my analytes of interest. I read that Phalanx is supposed to be resistant to high pH...don't have any experience myself with that Higgins product, but running an eluent at pH 9 or so can be a scary proposition that requires the use of a sacrificial column...silica begins dissolving at pH 8 or so.
Anyhow, what I'd try is to run your separation again, but at a more acidic pH...perhaps the imidazole-ion-pair may be better-retained than it is now and the other analyte less-retained...this may work to your advantage in a couple of ways in the long term.
I wish you well, don't know what the others will say, and these are only opinions.
MattM
Just an addition to Matt's valuable comments: The baseline drift you're observing is most probably due to your methanol gradient. At 220nm methanol shows significant absorption, so you will clearly see the absorption differences during your gradient run.
Is it really necessary to use a ternary gradient (buffer + methanol + acetonitrile)? At this wavelength, I'd throw out the methanol and run a gradient with buffer/acetonitrile alone. This will at least minimize the baseline shift. If you definitively need some amount of methanol to get the desired selectivity, I'd keep the methanol content constant and run a gradient in acetonitrile only.
Hi! Matt,
Thanks for your suggestions. I did get repeatable retention time for Imidazole though.Peak shape is not too good at pH 7.0 , see at tailing at end of peak. I will try to work around pH 2.5-3.0. My API pka is somewhere between 3.4-3.8 I think so I am not sure if the pH will work for it.
Will work on baseline drift today.
Monika
At lower pH imidazole is going to be protonated and you are not going to have too much retention, since pKa of imidazole is around 7. Right now you are in transitional state of imidazole. If you want to have it in non-ionized state you need to go up in ph at least to 9.
You answer is mixed-mode :)
Are you in US or outside of US? If you send me a sample I can show you how easy it is to have a short isocratic method at low UV or with LC/MS compatible conditions.
Vlad Orlovsky
www.helixchrom.com
Hi Again, bobby12,
Well...if the API has a pKa of 3.4 - 3.8 and imidazole has a pKa of 6.9...maybe acetate isn't such a bad idea after all. The imidazole will be protonated at pH 4.8 - 5.0 which would still work well with the octanesulfonate, and the API will be ionized and will elute more quickly, maybe such that you can switch to an isocratic elution if fortune holds. If both the API and imidazole are protonated and pair up with octanesulfonate, that could exacerbate the long retention (and need of gradient) for the API. The acetate may help a bit with the baseline drift, too, since it's cutoff is around 230 nm...
HPLCAddict explains things much better than I do a lot of the time...if MeOH can be avoided, it is a good idea. Alson, sometimes putting around 50 microliters of conc. HNO3 into the aqueous buffer per liter can help with drift in a gradient program.
Vlad also has a point...if only it was easy to always convince the bosses to have a variety of stationary phases to experiment with in the lab...I've had jobs where I couldn't switch stationary phases and those where it was no trouble at all to do so. If only we could always have that freedom and time to do things like that. The mixed-mode stationary phase could help retain the imidazole without an ion-pairing agent while not holding on so tightly to the API.
Best Wishes!
MattM