Importance of Controlled Room Temperature
For a UPLC-MS/MS system, how important is it that the room temperature is controlled to within a +/-5C range? If room temperature exceeds this range, what can an analyst expect to see in their results?
Some additional info about the UPLC-MS/MS system - gradient mobile phase, column at 40C, chemical residue analysis, ... let me know if you need additional info. I mainly just want an open discussion on the importance (or lack therof) of temperature control for a UPLC MS/MS system.
BHolmes
Any problem worthy of attack, proves its worth by hitting back...never give up!
Nowadays, controlled room temperature is not really important for the UPLC part. Most, if not every, UHPLC systems have a column oven (keeping the column temperature stable, and so the retention times) and a thermostated autosampler (keeping the samples temperature stable).
MS systems require a stable temperature to prevent any calibration shift, that could lead to mass precision deviation during the analyses.
If you use a high resolution mass spectrometer, you really should have a controlled room temperature within a +/-2C range (and prevent any rapid temperature fluctuation, even inside this range).
Also, even if your room temperature is controlled, never install a mass spectrometer near a heat source (or in front of a window without blinds, exposed to direct sunlight), or near the air conditioning system.
And in any case, always prevent temperature exceeding 30°C in the lab to prevent any overheating of the mass spectrometer.