LC-MS method for amoxicillin with sunfire HPLC colunm
Dear all,
I am trying to replicate a method for the analysis of amoxicillin on a HPLC-MS. The compound is diluted in water and the mobile phase is A:0.1% of formic acid in water and B: ACN. The gradient starts with 5% of ACN and it increases and the flow is 0,250 ml/min and it increases up to 100% of ACN in 12 min.
The column temperature is set up at 40ºC and I can’t control the temperature of the autosampler.
The peak shape are horrible and I am not able to get reproducibility. I am only able to get correct the first inyection.
Any idea or suggestion??
Thanks in advance
We need more information to help you.
What column do you use ?
When you say you're no able to get reproducibility, what do you mean ? Different retention time at each analysis ?
Do you let enough time after the end of the analysis for the column to get back to its initial state ?
The column is in the title it is a sunfire from waters.
I mean that after three inyection the compounds disapear..Is that possible?
What type of SunFire column do you use (C8, C18 ?...)
What are its dimensions ?
What is the volume of injection ?
Do you have an example of chromatogram ?
When you say the compound disappear after 3 injection, does that mean the 3 injections are identical and then you can't detect anything ?
Do you optimise the amoxicillin, i.e. infused directly to MS and get the optimum parameters?
Do you know that amoxicillin is ionised in negative mode?
In total there are 49 users online :: 1 registered, 0 hidden and 48 guests (based on users active over the past 5 minutes)
Users browsing this forum: Ahrefs [Bot] and 48 guests