Tablets: dependence of peak height/area on crushing or not
I have a tablet with two well-distributed APIs.
When I dissolve the entire tablet as it is, I get higher heights/areas for API 1 than when I crush the tablet before dissolving
At the same time, I get lower heights/areas for API 2 for non-crushed tablets compared to the crushed ones.
Does anybody knows such phenomenons as well as a solution?
Regards
Florian
Sounds like you have both a solubility issue with one API and a sticking problem with the other.
I'd be inclined to find a LC-friendly solvent capable of fully dissolving both APIs in the tablet matrix and avoid crushing the tablets. Use a number of whole tablets and a larger vessel for your sample prep.
DR
I'd do what DR suggests.
Why crush tablets if they dissolve? Extra step, chance of cross contamination, volatiles loss, etc.
In total there are 14 users online :: 2 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)