UV spectrum background problem
30 posts Page 1 of 2
Hello, I would like to describe a rather difficult to solve UV background problem that I could use some advise on. First I will describe the system we are using, then the problem we are having, and the attempts I have made to troubleshoot the problem.
System- Thermo Accela autosampler, 600 pump, and PDA detector. Column is C18 material and the mobile phase consists of methanol / water with 0.1% formic acid. We use gradient elution. The samples we are analyzing with this instrument are mostly plant extracts with relatively non-polar phenolic type compounds. We use a number of UV wavelengths depending on the compound but mostly 228 nm and 270 nm. The methods and samples have been used successfully in the past on other HPLC systems (Agilent for example). I am a bit new to Thermo equipment. The PDA detector was used but the rest of the equipment was new.
Problem- During gradient elution as the concentration of methanol (0.1% formic acid) increases the baseline goes up as normal but when one examines the UV spectra (200-400 nm) there is a peak with a maximum at ~236-237 nm and what looks like noise at lower wavelengths at beginning of run which eventually smooths out as another maximum at ~217 nm towards end of run. The absorbance of this peak goes up and up as the methanol (0.1% formic acid) increases. The problem with this background peak is it that it is close to 228 nm which I prefer to use for quantification of most of the compounds I am dealing with. It also makes it impossible to get a clean UV spectrum for peak purity analysis / identification because the background peak is large and gets in the way of seeing normal UV spectrum of compounds. This problem occurs in every sample I inject on this system whether its a blank, standard, or sample.
Troubleshooting- My first inclination was that this problem was an issue with contamination. So I set about a series of troubleshooting procedures to see if this was the issue and if I could solve it.
1- New mobile phases. No difference. I used different stocks of methanol and water which didn't make a difference. The formic acid is LC/MS grade and has been stored for about a year at 4 C. I have not yet tried a new batch of formic acid.
2- Flow individual solvents through PDA detector. Pure H20 = no background peak. Pure MeOH = no background peak. Pure H20 + 0.1% formic acid = no background peak. 50% MeOH with H20 = no background peak. MeOH + 0.1% formic acid = background peak. If I zero detector the peak will go away as expected but it always comes back if you start a gradient after zeroing.
3- Test each solvent channel in case there is contamination in line normally used for MeOH (0.1% formic acid). No difference the peak always appears with MeOH (0.1% formic acid).
4- Remove / replace column. New column or no column made no difference the peak is still there with MeOH (0.1% formic acid).
5- Replace UV and visible lamps. Lamps passed all usual calibration tests. No difference peak still there.
6- Replace solvent inlet filters. No difference peak still there.
7- Flush system with isopropyl alcohol to remove less polar impurities. No difference peak still there.
8- Flush system with 20% nitric acid in H2O. No difference peak still there.
After all this troubleshooting I am rather stumped. There are a few more possibilities left to try:
1- Get new formic acid. I find it hard to believe that LC/MS grade formic acid stored properly would be causing this problem. Has anyone else experienced problems like this from formic acid?
2- Replace PDA detector flow cell. This is an expensive part to replace and if it doesn't solve the problem it would be a total waste of money so I am reluctant to try this unless I am sure its a flow cell problem. Again the PDA detector was used and the problem has been with the system from day 1 it seems so that could be the problem.
3- Use different type of UV lamp. We using a CTS-A123 TSP UV 6000 lamp assembly. For some strange reason these lamp assemblies don't have the proper wire connections to connect directly to Thermoe Accela PDA detector and had to be rewired to connect. The reason I bring this up is I wonder if this wiring is part of the problem? I can't see how and I would expect if wiring was a problem the lamp wouldn't even turn on. I just found it weird that a lamp is sold as being a lamp for Thermo equipment yet doesn't directly connect to the detector without some rewiring.
4- Software problem? I again find this hard to believe but perhaps there is some setting that is not correct? Thermo software does have some detector settings I am a bit unfamiliar with such as rise time but I don't see how this could be an issue. But at this point I consider anything.
5- Pump contamination problem? I would have imagined nitric acid cleaning would have removed any organic impurities. I have been reluctant to take apart the pump and look for signs of contamination or breakage because everything else about this pump is pretty normal and I don't want to risk damaging the pump. Even if pump was problem I am unsure which parts would require replacement first. Although we do get occasional solenoid errors but I don't see how this could be connected?
Any other ideas? This is by far the most difficult and strangest problem I've ever dealt with on an HPLC. Any advise would be greatly appreciated.
Mmmh, strange, but even the system is new I would rinse the whole autosampler system, especially the injection system.
Was the column new? Maybe something sticks to the frit on the inlet.
Is it possible, that you produce Methyl formate in your solvent bottle?
If you are not bound on formic acid you could try TFA as acid.
You have already done a lot of experimentation. However, I would still suggest to go through the review
http://www.sciencedirect.com/science/ar ... 7304014657
"Ghost peaks in reversed-phase gradient HPLC: a review and update"
It could be a contaminant in formic acid. No matter how pure a reagent is purchased if it used by multiple users chances of contamination increase with time.
M. Farooq Wahab
mwahab@ualberta.ca
Thanks for help and replies.
Column was new and the autosampler was rinsed with the same cleaning procedures used for flow cell (IPA and 20% nitric acid).
Switching to another buffer is possible but not ideal since the method was validated with formic acid.
I've now tried two batches of formic acid and same problem. Although I still have not tried a completely new bottle yet.
Thanks for the article I'll give it a read and see if there is something I may have missed.
Agree with M. Farooq. Hard not to, he's a bright fellow.
Apologize in advance for what may be a poor question. Did your 20% nitric acid flush extend to the flowcell itself (and the flowcell may be made up of a material prohibiting the use of such a strong acid, anyway) and the degasser?
I've had Large problems with contaminated degassers in my past, such that I've taken the degasser out of the flow path to see what effect it has in troubleshooting I've done.
Please, see what you think, and thank you.
Oh...rise time refers to the time it takes for a signal...well, more like 67% of the signal, actually...sent from the detector to be transduced by the chromatography data system you're using. Important to set this value low for fast (narrow) peaks...Mac-Mod Analytical in the U.S. (and Hichrom in the U.K.) have published excellent guides on the proper use of this parameter.
MattM
Pretty good and systematic troubleshooting I'd say :D .
I think you nailed it down to the formic acid in methanol. From what you've described, I would rule out a contamination issue of the system...
Were these systems also equipped with PDAs or only single-wavelength detectors? Are you sure that this background peak never had been there in the past?
The absorbance of this peak goes up and up as the methanol (0.1% formic acid) increases.).
Unfortunately I'm not a spectroscopy expert, so I'm wondering what the UV absorption spectrum of formic acid in pure methanol is...I'm quite familiar with the carbonyl absorption in the range of, let's say 200-220 nm. Could it be that in pure methanol the absorption is shifted upwards due to different hydrogen-bonding properties or suppresion of ionization or whatever? So that the background peak you're observing actually is formic acid?
HPLCAddict's a smart guy, too...
The absorbance of formic acid in solution increases DRAMATICALLY at wavelengths lower than 230 nm, just like a mountain rising up from a plain. Don't know that this is as dramatic a case as the TFA-ACN charge-transfer complex, but HPLCAddict seems to be "on" to something here.
Source: p. 881 of Introduction to Modern Liquid Chromatography, 3rd edition, Synder, Kirkland and Dolan.
Someday, when I grow up a bit, if I could be half as smart as Misters Farooq, "Addict", that Imh guy and Peter what's-his-name that I can't remember at the moment...let alone Mr. Jupille...
MattM
Please do blank injections and prepare the blank just as you prepared the samples. If you see this ghost peak in blank, you know that you don't need to worry about it.
If not, then still there is nothing to worry about.
It is possible that your detector is too sensitive, change the spectral settings (read in steps of 2 nm instead of 1nm). This should help a bit.
As far as peak purity is concerned, I figured out that this is actually mathamatical juggelary. May be you want to try evaluating if you are getting sufficient NTPs for the peak of your interest ?
:D Thanks again for all your replies. I will respond individually.
SGA- Blank injections have the same issue. I don't know if calling this background problem a ghost peak is appropriate because its not really a peak its a consistent background in the UV spectrum. Either way you are right its not really a huge problem in that our results are good otherwise (reproducible accurate etc). The real problem with this background peak is it gets in the way of us seeing a clean UV spectrum. Plus I can't stand a HPLC problem that I don't understand so I have a desire to solve it.
Your idea about trying a wider nm slit is a very good idea, I will try that next time I calibrate the instrument and see if it makes a difference. I used to use 4nm on agilent HPLC and for whatever reason I tried 1nm on this thermo PDA. Could be the reason!
Mattmullaney- Yes there actually is a system that uses ammonium formate instead of formic acid that I could try. Its well validated method so that might be a good alternative to formic acid system. Not sure 100% if it will remove background problem but the method has other advantages in terms of resolution. I'd rather not work with TFA as we eventually want to incorporate this method into our LC-MS.
With regards to your questions about nitric acid cleaning. Yes I bypassed the degasser because Thermo said not to flow nitric acid through the degasser. They do recommend in their manual however, to use nitric acid as a final attempt to clean out the flow cell. However you are right though that this does leave the degasser as a potential source of contamination. Next time there is a gap in the instruments usage I will try bypassing the degasser and see if the background problem goes away.
Do you know where I can find that information on the rise time?
HPLCaddict- Yep the previous agilent systems I've used were PDA's and this background peak was never there. Thats why I thought its either contamination or me not using the hardware correctly. I am having trouble finding info on UV spectrum of formic acid in methanol.
Thanks again will post update after I've tried some other things to solve the problem!
Might be a bit difficult for me to find the info for the Accela...for the Waters and Agilent systems, this info is much more readily available. Please, stay tuned for advice on the hardware.
As to one of the Mac-Mod .pdfs, here it is:
http://www.mac-mod.com/pdf/technical-re ... uceECV.pdf
The info lies in Table 4, and is generally applicable. Don't be put off by the superficially-porous stuff. Also, time constant = rise time = peak width, depending on the CDS or instrument settings.
MattM
" The real problem with this background peak is it gets in the way of us seeing a clean UV spectrum."
Hmm...how is that possible ?
Try to increase the detection λ, this peak should completely disappear !
Regards.
It's not a peak in the chromatogram! It's a peak in the background UV-spectrum seen on the PDA.
Hmm...can you post a picture of what this looks like ?
Beginning of chromatogram where detector has been zeroed and UV background problem is not apparent.
Later in chromatogram where UV background problem is very obvious.
Note that this is a methanol blank sample.
yay I finally figured out how to post images :alien: