Transferring Loop injection to Syringe injection with split
2 posts Page 1 of 1
I have a vendor's GC headspace method for determination of impurities and have to adopt in my laboratory. The method uses a loop injection method while we a syringe injection method with a split mode option. How do I set the parameters to achieve the identical sensitivity, precision and accuracy?
It is completely impossible to get IDENTICAL sensitivity, precision and accuracy.
Start by calculating the actual volume transferred and match the incubation conditions and start from there.
2 posts Page 1 of 1