Sugar Separation with Boric Acid
Background: I am working on a Glucose and Fructose derivative separation. We purchased a Shodex Asahipak NH2P-50 based on a few papers that use the column. We are getting some separation, but it is not baseline resolved (MS detection). I noted in the Shodex literature that their Pb+ column may be better suited. Our current setup is with an LC-DAD-MS and we are getting an RID to see if we can sharpen up the peaks and improve separation.
My Question: I found this article (Journal of Chromatography A, 1101 (2006) 115–121) where boric acid is added to the mobile phase at a basic pH to complex to the sugar and enhance the separation while using an Aminex HPX-87K column. I was thinking of trying phenylboronic acid to try to gain the separation and perhaps get a UV signal. I would continue to use my Shodex column. Would my phenylboronic acid destroy my column by bonding to the amines? Is this a stable derivative? Any other suggestions to try are welcome.
All mentioned columns are excellent columns! Separating sugars in different matrices with different instrument setups is always challenging.
Therefore it is highly recommended to check All available application books AND send an application request to ALL the possible manufacturer, to their technical support team, BEFORE purchasing a column. That makes your live more easy.
I would discourage you from using Pb+ or boronic acid on any MS equipment. As far as the RI detector is concerned, I would ditch it and the Asahipak column and go for an amide HILIC column and ELSD. Advantages: ELSD is more sensitive than RI, can use the buffers used for MS and can work in gradient mode. For the column you could try a Waters Xbridge amide. You can see the applications notebook here http://www.waters.com/webassets/cms/lib ... 3438en.pdf. We use with satisfaction for the analysis of sugars in ELSD and MS. An extra advantage of this particular chemistry is the availability of many particle sizes. Given the low viscosity of HILIC mixtures, UPLC particle diameters cna be used with satisfaction on HPLC pumps, provided that the detector is fast enough.
FYI,
http://www.nacalai.co.jp/global/update/ ... _hplc.html
MeCN/H2O mobile phase.....ELSD.....