second level calib. for only one component in 6890 RGA
Hello everybody,
I am working on a 6890 RGA. Is it possible to do a 2nd level calibration for hydrogen alone? Because, I have only one calib. gas mixture and the hydrogen conc. in it is 80%. I get an error of 2 to 3 percent while using this calib. table. So I want to use 100% hydrogen as another standard, since my samples have conc. around 90 to 95 mole percent.
Please give me your suggestions, thank you all!
Dependent on your software, adding more levels would help with H2. I run Turbochrome and use 30,50,70, & 90% cal levels. Also dependent on the RGA, the hydrogen response may not be linear. Therefore necessary to have more levels. I have used 100% hydrogen as a check of the injection system, but since that wasn't a "calibration standard", I didn't use it as one. But I don't see any problem in using it. My RGA system from AC controls measures propane thru pentane components on FID. These are 1 level cals with a 3% std. The system has a fairly linear response on TCD for H2 from 2% to 90%. Now the Perkin Elmer is a different animal. The TCD hydrogen is non-linear.
If you are using bottled hydrogen and have a purity certification from the supplier I would believe that qualifies the cylinder contents to be used as a standard, although I should require a test of that particular cylinder ALONE, not merely a certification of the bulk gas from which the cylinder (bottle) was filled, to be a standard, even if it is 99.99 % hydrogen.
I would also not use the last 10-20% of the cylinder contents as a std.
Thank you very much, Les & chromatographer1 for your inputs.
I am using a Agilent 6890 RGA and I am not getting my methane peak. I am getting 79% N2 and 20% O2 (satisfactory separation) while injecting air and so I think the mol.sieve column is OK. I think there might be some problem in my valve switch-over timings. Can I use FID to detect methane? Since O2, N2 are not going to be detected on FID and I dont have CO in my sample, the peak I get after C5/C6+ peak, I assume, to be methane alone. And I am thinking about calibrating for methane using that peak from FID (and also the next peak in FID I am thinking to use for calibrating C2, since I dont get ethylene in my samples)
Am I right in doing this?
Your comments please.