qualification hplc agilent
4 posts Page 1 of 1
hi everyone :oops:
who make a qualification of Agilent HPLC 1200/ 1100
who can help me to handle OQ / PV programe ChemStation
I have the method and sequence but I CAN'T start the test and determine the limits
it's urgent please , where I can find a link or document
thank you
Hi.
I paraphrase your question:
Who can perform the qualification IQ/OQ for Agilent system, to meet the manufacturer's specifications, using Chemstation?
Of course, Agilent service engineers can help. Give them a call.
The qualification task will be more difficult than you think. (I have been there). Please consider to get help.
Hello
If you have Chemstation with "Verification" tab you can do it on your own. All you need is:
-certified flowmeter (or volumentric flask and watch)
-certified thermometer
-standards for UV/RI/FLD detector
Regards
Tomasz Kubowicz
Hello everyone,
I have an Agilent 1100 system with Chemstation B.04.03 and I want to do an OQ/PV but the methods and sequences are not included, where could I get them?
Thank you very much and greetings