Agilent 1260 RID Temp problem
hello,
get problem w/ the RID detector (Agilent G1362A), once the temp goes up by switching the method, it is hard to lower the temp down. ( it takes more than 3hrs to drop down from 35->30c ) do I suppose to turn off heater at this time (tried but no use) or keep the method setting of "Optical Unit temp is 30C", or it is just a HW issue for RID Temp controller? thank you!
That's a good question. I don't know if that RID has any cooling capacity; if it doesn't, and being well-insulated, just might take that long to drop temperature.
Our 1050 column heater is like that: can heat, but can only cool through heat loss.
Its normal. It doesn't have peltier element as JetStream column heater for example
Kind regards,
Ivan Styskin
LCCO Biotech
seems a HW issue, after changing the RID fan, it works (and can drop down to certain lower degrees!) but only hold for two weeks! doubt the board probelm, need to call service again!
Users browsing this forum: DuckDuckGo [Bot] and 2 guests
