Good morning !
We are currently analyzing pesticides (triazines, urones) in ground water using RP-LC-DAD. Mobile phase: water/acetonitril. Until now we just used MilliQ water from a water purification system (Millipore Elix) without any buffers. Unfortunately we observe retention time differences during a sequence run. Would you suggust us to buffer the water phase ? Which buffer would you suggest ? Do we have to add buffered water also to acetonitrile ?
I very appreciate your suggestions. Thanks
Daniel
Any time that ionization of either the column or the analyte is an influence on the retention time, you should use a buffer. There have been many discussions of buffers on this Forum, and I recommend you look through some of the old postings first.
Mark Tracy
Senior Chemist
Dionex Corp.
Greetings,
You are analyzing some organic molecules that will probably not appreciable ionize in and aqueous/organic solvent mixture. I would doubt that the absence of a buffer is causing your retention time issue, especially since you seem to state that you have some history of successful analyses with pure water. I would suspect more simple issues like pump or proportioning system failure.
When you hear hoofbeats, look for horses, not zebra.
Evan L. Cooper