Problem with a GC MS from ThermoFisher - Automatic EI tuning
2 posts Page 1 of 1
Good morning,
I'm everything but a specialist in GC MS but recently I had a lot of problem with it...
I have many errors when I perform an autotune...
I'm attaching the image and I would be glad if anyone could help.
Thanks a lot,
Kevin
Here is the manual. GTFY
http://www.thermoscientific.com/content ... W-RevB.pdf
Compased to the example tune, the intensities on this instrument are extremely low. I would start by trying to figure out why that is. Most likely dirty source or calibration compound is running out. Have you checked for leaks?
2 posts Page 1 of 1