DIONEX P680 PUMP
i HAVE A PROBLEM WITH THE DEGASSER, I WANT TO KNOW HOW TO OPEN THE MODULE.
If you send an email to stefandotbrandatthermofisherdotcom i will send you the service instruction manual with detailled information how to change the degasser (and how to open the module...).
Good luck
Good luck
Stefan Brand
Thermo Fisher Scientific
Does anyone have experience with a Dionex P680 LPG pump. the proportional valve is not operating at all. It appears that no voltages are applied to the solenoid valves. I am thinking that the board the valve is connected to might be faulty? Any suggestions?
Dan
Dan
I'm having a similar issue with one of our P680 pumps. Is there a way to examine the degasser? I'm guessing there is a leak in the vacuum system.
If you have Dionex pump with the integrated 4-channel degasser module (e.g. P680A LPG-4), then you can remove the degasser module component assembly (vac pump with 4 chambers and tubing plus PCB) and send the module out for professional diagnostic testing and repair [http://www.chiralizer.com/uploads/8/9/7/7/89770565/hplc_degasser_repair_services_page.pdf]. The HPLC vacuum degasser should be professionally cleaned and serviced every 4 to 5 years (their max lifetime is 4 to 5 years). A worn or damaged vacuum degasser can contaminate your mobile phase so it is important to have them professionally serviced at regular intervals, or whenever a problem is observed.
The Pump channel solenoid valves only receive DC voltage when a channel is selected via the controller (very fast).