HPLC with DMSO as solvent
Is it possible to use HPLC to analyze the concentration of small molecules dissolved in DMSO? If not, is there a suitable method?
We use DMF and DMSO for dissolving samples for LC-MS analyses.
As long as you keep the injection volume low, 5-10 ul's, causes no serious problems and still gives decent peak shapes.
Sailor
DMSO is used as surfactant to solubilise sample in HPLC. Being a strong solvent should be used in less amount.Is it possible to use HPLC to analyze the concentration of small molecules dissolved in DMSO? If not, is there a suitable method?
Amaryl.
We have used DMSO (water-free) as diluent for a substance that very sensitive to water. The sample is injected in less than 20 uL and found no significant peak distortion.
syx wrote:
Was it with Normal-phase or Reversed-phase mode?We have used DMSO (water-free) as diluent for a substance that very sensitive to water. The sample is injected in less than 20 uL and found no significant peak distortion.
RP-LC. The mobile phase contains buffer solution. However, the aqueous content of the mobile phase did not degrade the active substance, because the analysis time is not more than 15 minutes.
Users browsing this forum: Semrush [Bot] and 9 guests