agilent 7890A sudden power off
9 posts Page 1 of 1
I have a problem that my Agilent 7890A GC had suddenly powered off and when i tried to turn it on it goes off after the self-checking and displaying the Agilent technologies logo. What can I do ,please?
A '' thermal shutdown '' error may cause this , possible causes : faulty cooling fan ( at the back of the injection port ), faulty temperature sensor or heating element.
Is Agilent logo the only information from GC display ?
Any information about Power On Selftest ?
Thank you guys for the quick support
this is what happens at the self-test Image
then the Agilent logo displayed then the device turns off
no other notifications are displayed
there is 2 red and 1 orange lights at the back of the device
I don't know what does this mean ?
I was hoping for a kind of error notification on the display.
Unfortunately selftest won't help with diagnosing.
so does this problem require changing the hole main board of the device ?
As mentioned in the other thread, can you please check your line voltage?
John
Did you find the problem,?
We have same error too and led on the back side turn on . It seems thermal shut down.
This problem will require the replacement of the GC mainboard.
9 posts Page 1 of 1