Why people use different TFA conc in MP A and MP B?
Hi, All,
I wonder why people use 0.05%TFA in H2O as MP A, while using 0.04% TFA in ACN as MP B? What is the rational behind using the different TFA conc. in MP A and MP B?
Your answers are appreciated.
if people use different TFA concs in some methods there might be a reason. The best thing is to ask the original authors.
Maybe at at specific wavelength the absorption is equal, thuse avoiding a "gradient ramp" in the chromatogramm.
Maybe someone thought about pH in a way that compares H+ and OH- ratios. As ACN is aprotic smaller amount of TFA could have same effect.
Alex
Wangge
I agree with Alex first comment - that they are probably using a low wavelength close to the cut-off for TFA to attempt to produce a straight baseline.
WK
wangge,
Alex was correct. The reason is to produce a flat baseline in the 210-220 nm range during the gradient. When I prepare mobile phases like this I use 0.1% TFA in 95:5 water/acetonitrile for A and 0.08% TFA in 5:95 water/acetonitrile for B. After preparing the mobile phases, I measure the absorbance of each mobile phase at 215 nm and add a few drops of TFA if necessary so that the absorbance of A is 0.03-0.04 AU greater than B.
The absorbance spectrum and extinction coefficient of TFA changes with the concentration of acetonitrile. At 214 nm, the shift is minimized. The different concentrations in A and B compensate for that. This trick works well as long as the gradient is slow, but sharp gradients interact with the column to give nonlinear behavior.
Mark Tracy
Senior Chemist
Dionex Corp.
In total there are 19 users online :: 2 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], John Guajardo and 17 guests
