Dear all,
Since i do not have practical experience with PTV injection, this might sound naive, but i was wondering if a PTV injector, in its hot split injection mode, leaving the split vent continuously open and introducing the sample into a hot inlet, can be used or seen as a classical split injector?
Thanks for your comments.
Best regards,
Kurt
Hi Kurt
Simple (simplistic !!) answer: Yes, but with all the foibles and idiosyncracies of an ordinary split-splitless.
Most (all ?) PTVs have smaller liners than the dedicated split inlets so there might be some issues of vapour volume and heat transfer.
Give it a try and let us know how it goes.
In my experience fairly volatile compounds (C20 or less) at a relatively high concentration and a high split ratio work best when using a PTV as a split injector. The liner volume is usually very small, so you need a high flow through the liner to minimize overloading of the liner.
Kurt,
The vapor volume of the injected sample can be estimated from a GC Pressure/Flow calculator software downloadable in Agilent's website. The calculator is fairly useful as a hint on how much volume you could inject in different size of liner at different column head pressure and inlet temperature.
http://www.chem.agilent.com/cag/servsup ... s/GCFC.htm
Hope this helps.
Kurt,
Hope this helps.
Yes, follow this advice. Kudos Chhubert!
This is a necessary tool for ALL chromatographers.