KPF6
I was asked to set up a method with this chemical, potassium hexafluorophosphate. I never used it before and it has a MSDS longer than most.
The method is a regular RP-HPLC method and I cannot really see why it is being used. What are the benefits of this buffer?
Hello
One of the LC method for basic compounds is to use chaotropic ion which can form neutral ion pair (hydrophylic in mobile phase). Common one is PF6-
As I remember the other one that works as well is ClO4-
Tomasz Kubowicz
Hello
Tomasz Kubowicz
Thanks Tomasz!
So it is a peak shape improver for basic compounds more or less
And PF6- (a chaotropic agent, as per the Hofmeister series of chaotropic and kosmotropic agents) will also generally increase retention of polar analytes to a degree.
If you're interested, look up papers written by Kazakevich and LoBrutto such as:
"Reversed-phase high-performance liquid chromatography behavior of chaotropic counteranions", Journal of Chromatography A, Volume 1064, Issue 1, 28 January 2005, Pages 9-18, by Y.V. Kazakevich, R. LoBrutto and R. Vivilecchia.
and
"Effect of counter-anion concentration on retention in high-performance liquid chromatography of protonated basic analytes", Journal of Chromatography A, Volume 913, Issues 1–2, 13 April 2001, Pages 189-196, by R. LoBrutto, A. Jones and Y.V. Kazakevich.
MattM
Users browsing this forum: Bing [Bot], Semrush [Bot] and 8 guests
