Internal Std varition in 1st injection
3 posts Page 1 of 1
Hi hi hi everybody
While doing the system suitability (MQC-Aqueous), the first injection Internal std area is very lesser than the other consecutive five injections.
So that my RSD is failing.
Always the first injection will be problem. Why?
But this problem not in the plasma samples. Why?
Anybody can help me?
My maid id lingaaexpress@gmail.com
lingaa
I never use my first injection. Never.
An attempt to explain: Some substances "find" some places on the stationary phase to which they bind strongly. Once these sites are covered with this substance the effect of essentially partially removing this substance from the chromatographic process will cease.
Now, I wonder what you mean with first time? The first time after washing or standing without flow......?
I can´t say that I have seen this thing with all analytes that were tried here. Even proteins, which have a tendency to do this, can be "tought" to behave if the proper mobile phase, stat phase is used. Of course, this is a matter of what you want to tolerate, if a certain amount of tailing is tolerated (or must be tolerated) one can also expect (and maybe tolerate) some lowering in early injections.
3 posts Page 1 of 1
In total there are 30 users online :: 0 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 30 guests