negative intercept and no peaks
hello every one . IM Still a beginner in lcmsms . I tried to measure my drug sunitinib samples in mice plasma and used an internal stranded verapamil 50nM and made a calibration standreds from 1nM to 3000nM but the problem is in some samples I GOT NO peaks and also the intercept was negative so please can you help me in addressing the problem
You'll have to post more details or we just have to guess.
It seems you have 2 different problems: a calibration problem and a sample preparation problem.
First get your calibration right. This should be the easy part.
I agree with RnDirk, but a negative intercept indicates that you might be hitting the S/N limit of your method.
Think about it this way: There needs to be a certain amount of material present to give a peak at all. All concentrations below that amount will give 0 peak area.
Also: 1 - 3000 nM is a very wide concentration range for LCMS. it is unlikely that you will have good linearity over such a a broad range. See if the response factors are equal over the range (area/concentration).
Good Luck!
Arne
Good Luck!
Arne
thank you very much for every one reply . Im sorry i tried to upload my excel file but i couldn't Im trying to learn as much as possible . thank you
In total there are 23 users online :: 2 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Semrush [Bot] and 21 guests