waters dual wavelength UV detector/PDA
:?: Are there any visible lamps in waters dual wavelength UV detector/PDA? How to use them at visible wavelength range, such as 525nm?
Thanks.
Hi CQ_32,
Yes, there are!
How to use them? You just enter the desired wavelength (525 nm) in the Instrument Method and this should do the trick. Remember though to save the changes :wink:
The problem is that there is no tungsten lamp installed in Waters UV detectors, such as 996 PDA and Dural Wavelength. A filter has to be used. Has anyone have this experience?
Those detectors use a duterium lamp which produces wavelengths well into the visible range (700 or 800 nm I think).
As AA mentioned there are deuterium lamps in those detectors and their wavelength range is from 190 to 700 nm. No filters needed. The grating will adjust the wavelength for you :wink:
Thank you for the information.
However, I did see the message " filter needs to be installed" for the wavelength I chose 525 nm for dual wavelength detector. Also there is a place inside the detector for filter.
For PDA 996, I didn't see such things. So I assume it is OK.
Hi CQ_32
I’ve just tried to enter 525 nm in Waters 2487 UV/VIS detector method and I didn’t get any massages. So it works fine for me. I use Waters Millennium software. If you use another vendor’s software, it just might be a “bugâ€
