Headspace 6890 and 7697 - no trigger to acquire
Sorry -I put this in wrong subject
Hi ,
Headspace 7697A carries out equilibration , pressurization and injection down transfer line
HP6890 then starts oven program (lights on HP6890 turn on)
Empower 3 sits on status 'Preparing - Sample for injection'
No trigger to collect and acquire data
Lace boxes all working and pinging (IP addresses ok) and you can monitor baseline and detector output ok
Any thoughts ? Thanks
Hi ,
Still no forward with this
When the 7697A headspace injects the status time on GC 6890 immediately starts counting upwards (Time = 0.xyz) so injection is recognised from HS to GC
Empower 3 remains on 'waiting on injection'
Any thoughts much appreciated
Hi ,
I did notice configuration on back ALS injector
What should this be set at ? Does this come into play with remote control ?
Current options
No Headspace
Headspace 7697A (12 Vial)
Headspace 7697A (111Vial)
But when I select Headspace 7697A (111Vial) the GC out put tries to communicate with Empower and counts up
Downloading method info from Empower to GC and HS not a problem
