TSQ Vantage - background in negative mode.
Hello all. In last time in negative mode appeared a huge background peak with 169 m/z. What can it be and how to remove it ? I've tried to clean ESi box, wash tubes but without effect.
Thanks!
Did you isolate it to the MS only?
In case calibration with polythyrosine (direct infusion in mass) or direct conection LC without column - the same strong 169 m/z.
What if you just turn on the MS without infusion or column flow?
If it's still there and it's bothering your analysis, i'd advice to take cleaning 1 step further and clean the part behind the transfer tube: the first lens system/ion funnel.
If you don't have experience doing that it's best to ask Thermo.
If it's still there and it's bothering your analysis...
I tried to do such way,but peak is still here (.
So i have to clean capillary and ion optics(we can do it without Thermo service).
Thank you so much!
Sorry for confusion. Ii checked out system one more time - contamination with 169 m/z goes from channel of pump with organic phase. Very strange thing - i got excellent tune with polythyrosine in positive and negative modes, but instrumental method in MRM doesn't work properly. MRM in negative - no signal. in positive -everything good.