if any body have a good HPLC method for cefdinir?
Dear dhru,
You could at least use a couple of seconds to finish the sentence, in return for all this time you expect the folks on this forum to invest in finding an analytical procedure as required :roll:
Dear Danko
i all ready tried on different pH but i am unable to get a good peak shape of cefdinir.i know that pH is not only the criteria to get good peak shape.
i check it by changing different column , organic solvents,diluent ,organic modifier , eventhough i didnt get good peak shape
People will be more willing to help you if you post all of the conditions of your method.
List the column, mobile phase, and sample concentrations and diluent.
Method development for the determination of cefdinir and its related substances by high-performance liquid chromatography.
Okamoto Y, Itoh K, Namiki Y, Matsushita J, Fujioka M, Yasuda T.
J Pharm Biomed Anal. 1996 Apr;14(6):739-48.
Isolation, structural elucidation and characterization of impurities in Cefdinir.Rao KV, Rani A, Reddy AV, Bharathi CH, Dandala R, Naidu A.
J Pharm Biomed Anal. 2006 Dec 14
Bruce Hamilton
Thank you very much
i got that method but still i have one question that can we use dioxane in our mp. they gave 20 ml in 800 ml
dhru