signal-to-noise: standard calculation procedure?
Hi everyone!
We are just discussing several different methods of calculating signal-to-noise ratios in chromatography (predominantly GC/MS).
baseline level = average value of local noise left and right of target peak inside 20*half-peak-width, while ignoring all data closer than 5*half-peak-width (20 and 5 = arbitrarily choosen)
noise =
a) difference between maximum and minimum value of local noise
b) standard deviation of local noise (a lot smaller!)
1) Do you know any formal requirements or standard procedures to calculate S/N ratios?
2) Do you determine the noise by method a or b? Or other? I know many people prefer standard deviation because it gives numerically better S/N. However, in spectroscopy (nmr) usually strict min-max values are used.
3) How large do you select the local noise range? How close, how far?
Obviously, the S/N is only valid for a particular acquisition method, because scan rate, scan range, dwell time and so on do influence the S/N ratio. Naturally, spikes are part of the noise. Only reproducible peaks may be ignored.
Thanks for your comments!
Kajjo
John W. Dolan, LC.GC Europe, January 2006, p. 12.
Thanks for the reference. However, my questions are not addressed at all in that article. It is just very basic.
Kajjo
In total there are 33 users online :: 0 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 33 guests