Minimum Detection Limit
3 posts Page 1 of 1
Hello does anyone have a method or procedure for calculating MDL's
You can find it in the USP, but I don't have the reference handy. The FDA has a "crude" way of estimating it on page 12 of their "Reviewer Guidance for Validation of Chromatographic Methods (http://www.fda.gov/cder/guidance/cmc3.pdf )
If you don't have access to the USP, you can select one of the methods defined in section 6 " Detection Limit " of ICH Q2(R1) which is called " Validation of Analytical Procedures : Methodology ", and is freely available from ICH web site.
If you're not in the pharmaceutical industry, various other regulatory authorities, such as FDA, define their preferred procedures. Often the information is freely available on web sites.
Bruce Hamilton
3 posts Page 1 of 1
In total there are 3 users online :: 1 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)