separation column for krypton and nitrogen
Let me know which column is the best for a good separation of Krypton and Nitrogen in Helium.
I use a Packed Mol. Sieve 5A 100/120 12' long and it works not so bad, but I would like to know if something better is available.
thank you
dany
if you have already the selectivity, in terms of resolution and efficiency you can use a PLOT (porous layer open tubular) Mol Sieve column.
philippe
Yes, Dany,
A 1-2 ft Supelco Carboxen 1000 1/8" packed column will work wonders.
A Supelco PLOT Carboxen 1006 will also work well but cost a lot more money.
Send me your email address and I can send you a chromatogram of a 2 ft column.
Rodney George
Gas Separations Research
Supelco
595 North Harrison Road
Bellefonte, PA 16823
814-359-5737 voice
814-359-5459 fax
rodney.george@sial.com
Dany,
Here are the specifics:
The separation of nitrogen and krypton is achieved on a 2ft Carboxen 1000 60/80 mesh column that is only 2 ft in length at 40°C: neon 0.092 minutes, nitrogen 0.390 minutes, krypton 1.847 minutes.
I also have an application that separates hydrogen, oxygen, nitrogen, and krypton in less than two minutes.
Rodney George
Supelco
814-359-5737 voice
814-359-5459 fax
