Do you have some suggests for Tadalafil impurities in USP ?
I'm following the procedure of USP about the Tadalafil organic impurities.
In the below control and situation:
Mode : LC
Wavelength : 285 nm
Flow rate : 1.0 mL/min
Injection volume : 20 micro-liter
Column : 4.6-mm X 25-cm; 5-micrometer packing L7
I don't know which peak is we-wanted.
Do you have any suggest or spectrum for me to referring?
Thanks.
Hello.
I wish USP & EP both provide harmonized monographs: https://extranet.edqm.eu/4DLink1/pdfs/c ... s/2606.pdf
I made a year ago raw material QC according to EP so spectra are also available if sill required.
Best regards,
Dmitriy A. Perlow
[Note— The relative retention times for tadalafil and the 6R ,12aS diastereomer of tadalafil are about 1.0 and 1.03, respectively. ]
from this i think tadalafil peak is the first peak and it diasteriomer is the second
Not very good chromatography. Capacity factor should be at least 2. What are your chromatographic condition? Do you have a resolution requirement in your system suitability acceptance criteria?
