Method validation problem, help needed
Dear all,
Do you think I need to redo all my validation studies for the case described below:-
Our laboratory developed a set of clean up procedures for the analysis of pesticide residues in vegetable oil matrix.
According to original method, the final sample is analysed with GC-PFPD, and few parameters were validated i.e. linearility, recovery, RSD%.
Recently, the detector was down and I switched the analysis to another GC-FPD (Caution: detector was different), I built a new calibration curve and found that the linearity was still there within the analytical range.
My question is:
1. Do I need to redo the recovery study?
I understand that the LOD and LOQ may be different in that two system, however the samples I am working on were roughly 4X higher than the lowest calibration concentration.
In my opinion, I think the changing of detector has no direct effect on the clean up procedures performance (correct me if I am wrong).
Can anyone submit your suggestions and views, thank you.
If any one of you agreed to my point of view, can you tell me where can I find the proof (journat, discussion paper...etc).
Thank you very much
Kenny
I think this is the wrong group to ask.
You need to ask the quality section of your organisation, the clients, and/or the relevant regulatory agency.
If I were in your situation, I would revalidate, just for my own peace of mind. Changing to a different detector is a significant modification that should be clearly documented, including a simple risk analysis to justify any decisions.
Please keep having fun,
Bruce Hamilton