Chromatographic Reports in Empower 2
Is there a way to create a chromatographic report that will give a chromatogram and peak table on individual pages for each separate injection in a sequence.
All I can seem to get is all the Chromatograms one after the other and then one peak table at the end.
Although I can essentially get what I want by using the 'Inidividual Report' wizard, this means I can only print off one sample at a time. Is that correct? or I'm I missing something?
When I use the 'Summary Report' wizard, all I seem to end up with is all my chromatograms one after another, followed by one large peak table. I have tried putting a page break after the table on the report creator, but to no avail..
Suggetions please!! :evil:
Stick a peak table and a page break under the chromatogram in the report editor. Click and drag a box over the lot of them and group them (^g). Then, right click in a "blank" patch of the group and select composite properties. Order the group by injection ID or whatever you wish. If you leave the overall method set to "Summary by all" (right click on a margin, outside any report object to get to method properties), you should get 1 page of plot & table for every injection.
The Multi Sample Summary report method (which is in the Defaults project) is a good place to start.
