negative baseline drift
I'm getting a negative baseline. I have changed the liner, septum, checked the column's connections. How can I fix this?
Try to maintain a positive attitude! Sorry, I couldn't help myself :) Please post more details so that we can have a clue about what the cause could be.
It sounds like the detector does not have the zero set properly. Please give more details, type of instrument, detector, etc.
It's an FID. 6890 GC. There is a temperature ramp.
Still not enought information! Give us as many details as possible about your system and analysis parameters. Have you double-checked everything (gas flows, detector signal, etc.)? It's always a good idea to have a written record of all parameters when the system is working well in order to compare when problems arise.
I took the column out, and put it back in, and i don't have the negative drift. But now I'm having another problem. My chromatogram looks like there's an indication of column bleed. I ran a column compensation and that didn't help. SHould i try a new column?
Are you seeing column bleed from the same column that you used before that did not exhibit bleed? If so, one possibility is oxygen getting in from a gas leak. You can certainly try a new column but in my experience, a new column may exhibit more bleed than an aged column. Check your connections and parameters again.