Hello.
Anybody know how i can create a custom report from "Report Layout" that includes USP Tailing, Resolution, Theoretical Plateau,? This informatios is only available in predefined reports, but i can not put it in "Report Layout" for my personalized reports.
Thanks.
Daniel
Anybody know how i can create a custom report from "Report Layout" that includes USP Tailing, Resolution, Theoretical Plateau,? This information is only available in predefined reports, but i can not put it in "Report Layout" for my personalized reports.
Yes, we did this routinely for our pharmaceutical assays, on A.07.01 and A.09.03, same process. We had the Chemstation do a Sequence Summary Report at the end of each sequence. See this document, should explain most of it. https://www.agilent.com/cs/library/supp ... faq205.pdf
I'm a computer idiot and I was able to set this up for my department, could also to RSDs of the calibration standards, RSD of the retention times, etc.